Posted
When there is an access problem affecting clearance, does it always show up in the arterial and venous pressures first?
 
Reply With QuoteEdit or Delete MessageReport This Post
<ole school>
Posted
No .One reason could be recirculation.Have your clinic do a recirculation test to see if this is the problem.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Pam N.>
Posted
Please define recirculation.
 
Reply With QuoteEdit or Delete MessageReport This Post
<ole school>
Posted
ACCESS RECIRCULATION is the measurement of the % of the whole blood that returns directly from the venous port to the arterial port without first passing through the heart and other vascular spaces of the body.Significant recirculation reduces the effectiveness of the Dialysis therapy. Hope this helps
 
Reply With QuoteEdit or Delete MessageReport This Post
<Pam N.>
Posted
Yes thanks. So then what causes recirculation, are there any telltale signs when recirc. is the problem, should anything else be ruled out (i.e high conductivity alarm keeps going off) before going to a procedure to check the access? What is that procedure called and what does it involve?
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
If the conductivity alarm keeps shutting the machine off on a day when your labs are being drawn, obviously it will affect your lab results-- and you should make sure your doctor is aware that it could be a machine problem before he/she sends you for a fistulagram.
 