December 24, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for December 24, 2003:
PRODUCT:
a) Mahurkar QPlus Dual Lumen Catheter Kit with Curved Extension, 13.5 Fr/Ch x 19.5cm For hemodialysis, apheresis, and infusion. Reference: 8888-135192.
Recall # Z-0233-04;
b) Mahurkar QPlus Dual Lumen Catheter Insertion Tray with Curved Extension, 13.5 Fr/Ch x 19.5cm For hemodialysis, apheresis, and infusion Reference: 8888-135195.
Recall # Z-0234-04.

This is a Class III recall - a situation in which use of or exposure to a violative product is not likely to cause adverse health consequences.

CODE
a) Lot Number: 140885;
b) Lot Number: 140774.

RECALLING FIRM/MANUFACTURER
Kendall Healthcare Products Co, Mansfield, MA, by letter on December 1, 2003. Firm initiated recall is ongoing.

REASON
Product is mislabeled to catheter length. Product labeled as a 13.5 Fr x 19.5 cm may contain a 13.5Fr x 16.0 cm.

VOLUME OF PRODUCT IN COMMERCE
55 catheters.

DISTRIBUTION
IL, MA, OH, TN, TX, VA.


December 18, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for December 17, 2003:
PRODUCT: CA Cellulose Acetate High Efficiency Hollow Fiber Dialyzers; Model CA 210, product code 5M1736; 24 units per case. Recall # Z-0177-04.

This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

CODE
Lot numbers A00K09, exp 10/2003; A01C04, exp 2/2004; C01A08, exp 12/2003; E01B15, exp 1/2004.

RECALLING FIRM/MANUFACTURER
Baxter Healthcare Corp., Round Lake, IL, by letters dated October 30, 2003. Firm initiated recall is ongoing.

REASON
Use of the dialyzers may cause iritis (red eye) patient reactions.

VOLUME OF PRODUCT IN COMMERCE
34,865 units.

DISTRIBUTION
Nationwide, and Internationally.


October 28, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for October 29, 2003:
PRODUCT: LifeSite Hemodialysis Cannula Exchange Kit
Part Number: LHCEK0000.
Recall Z-0045-04.

This is a Class III recall - a situation in which use of or exposure to a violative product is not likely to cause adverse health consequences.

CODE
Lot Number: 30956.

RECALLING FIRM/MANUFACTURER
Vasca, Inc., Tewksbury, MA, by letter on September 22, 2003. Firm initiated recall is ongoing.

REASON
Outer kit mislabeled with an extended expiration date.

VOLUME OF PRODUCT IN COMMERCE
81 units.

DISTRIBUTION
Nationwide.


September 11, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for September 10, 2003:
PRODUCT: (Various) Baxter Solution Sets (see report)

This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

RECALLING FIRM/MANUFACTURER
Baxter Healthcare, Corp., Round Lake, IL, by letter dated July 16, 2003. Firm initiated recall is ongoing.

REASON
There are possible pinholes or perforations in the set tubing, caused by damage that occurred during the packaging process.

VOLUME OF PRODUCT IN COMMERCE
4,534,365 units.

DISTRIBUTION
Nationwide and Internationally.


July 24, 2003 - There is two recent medical product recalls of interest to the dialysis community in the FDA Enforcement Report for July 23, 2003:
PRODUCT: Nipro SafeTouch Safety Fistula Needle, Baxter Safety AV Fistula Set and EXEL Secure Touch Safety AV Fistula Needle Set. Recall # Z-0408-03.

This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

CODE
All catalog numbers and all lot numbers distributed between 10/24/2001 and 11/01/2002.

RECALLING FIRM/MANUFACTURER
Nipro Corp, Miami, FL, by letter on November 8, 2002. Firm initiated recall is ongoing.

REASON
Air or fluid leakage at the hub/needle junction.

VOLUME OF PRODUCT IN COMMERCE
6,984,000.

DISTRIBUTION
Nationwide and Canada.


PRODUCT: Various IV adminstration sets and tubing. See the FDA Enforcement Report for a complete listing.

This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

RECALLING FIRM/MANUFACTURER
Abbott Laboratories HPD/ADD, Abbott Park, IL, by letters dated June 9, 2003, June 13, 2003, and July 7, 2003. Firm initiated recall is ongoing.

REASON
The drip chambers and/or adapters in the sets were manufactured with non-radiation grade plastic resin, resulting in discoloration of the resin.

VOLUME OF PRODUCT IN COMMERCE
70,062,517 sets, 1,079,350 chambers.

DISTRIBUTION
Nationwide and Internationally.


May 29, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for May 28, 2003:
PRODUCT:
Baxter System 1000 Single Patient Hemodialysis Delivery System; Model SYS1000; all series 1000 instruments labeled as System 1000, AltraTouch 1000, Baxter Tina and Baxter Aurora.


Recall # Z-0833-03.

This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

CODE
Serial numbers 1001S to 1142S; 50001 to 52109; 01001 to 23764.

RECALLING FIRM/MANUFACTURER
Baxter Healthcare Corp., Round Lake, IL, by letters on March 20, 2003. Firm initiated recall is ongoing.

REASON
The dialysis treatment may be run without the bicarbonate due to an undetected bicarbonate pump stall following a sodium change.

VOLUME OF PRODUCT IN COMMERCE
25,014 units.

DISTRIBUTION
Nationwide and Internationally.


May 28, 2003 - Baxter International Inc. last month received a warning from U.S. regulators for allegedly failing to meet quality control standards in the production of an intravenous tubing system, a letter released Tuesday said. Story from Reuters/Yahoo. (link is no longer available)

The FDA posted the letter to Baxter on its web site (pdf format).

May 8, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for May 7, 2003:
PRODUCT:
a) Tandem-Cath Hemodialysis Cuffed Catheter with side holes

10 Fr 19 cm Arterial Cuff/22cm Venous Cuff Code #: 8888-219226. Recall # Z-0742-03;
b) Tandem-Cath Hemodialysis Cuffed Catheter with side holes
10 Fr 23 cm Arterial Cuff/26cm Venous Cuff Code #: 8888-223266. Recall # Z-0743-03;
c) Tandem-Cath Hemodialysis Cuffed Catheter with side holes
10 Fr 28 cm Arterial Cuff/31cm Venous Cuff Code #: 8888-228316. Recall # Z-0744-03;
d) Tandem-Cath Hemodialysis Cuffed Catheter with no side holes
10 Fr 19 cm Arterial Cuff/22cm Venous Cuff Code #:8888-219220. Recall # Z-0745-03;
e) Tandem-Cath Hemodialysis Cuffed Catheter with no side holes
10 Fr 23 cm Arterial Cuff/26 cm Venous Cuff Code #: 8888-223260. Recall # Z-0746-03;
f) Tandem-Cath Hemodialysis Cuffed Catheter with no side holes
10 Fr 28 cm Arterial Cuff/31cm Venous Cuff Code #:8888-228310 Recall # Z-0747-03.

This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

CODE
All codes.

RECALLING FIRM/MANUFACTURER
Tyco Healthcare/Kendall, Mansfield, MA, by letter on March 20, 2003. Firm initiated recall is ongoing.

REASON
Tandem Hemodialysis Catheter may have splits in the catheter shaft below the extension adapter.

VOLUME OF PRODUCT IN COMMERCE
1,077 units.

DISTRIBUTION
Nationwide and Internationally.


April 11, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for April 9, 2003:
PRODUCT
a) Baxter SPS 550 Single Patient System Hemodialysis Machine. Recall # Z-0660-03;

b) Baxter 1550 Single Patient System Hemodialysis Machines.
Recall # Z-0661-03.

CODE
a) Catalog numbers: 5M1309, 5M1309R, 5M1310, 5M1310R, 5M5506, 5M5506R, 5M5507, 5M5507R, 5M5516, 5M5516R, 5M5518, 5M5518R, 5M5533, 5M5536, FM4649, FM4654, FM4714, FM4796R, FM4797; all serial numbers;

b) Catalog numbers: 5M5538, 5M5538R, 5M5551, 5M5551R, 5M5575, FM4719, FM4774, FM4779, FM4779R; all serial numbers.

This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

RECALLING FIRM/MANUFACTURER
Recalling Firm: Baxter Healthcare Corp., Round Lake, IL, by letters dated February 20, 2003.
Manufacturer: Baxter Healthcare Corp., Deerfield, IL. Firm initiated recall is ongoing.

REASON
Excess or insufficient ultrafiltration may occur during manual control of ultrafiltration via adjustment of transmembrane pressure (TMP).

VOLUME OF PRODUCT IN COMMERCE
17,322 units.

DISTRIBUTION
Nationwide and Internationally.


March 13, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for March 12, 2003:
    PRODUCT: Baxter Meridian Hemodialysis Instrument.

    The instrument is part of a high permeability hemodialysis system which consists of a controlled dialysate delivery system that incorporates an ultra filtration controller to prevent excessive loss of water from the patient's blood, an extra corporeal blood set, and a high permeability dialyzer. The standard features of the Meridian instrument include a high blood flow rate capacity (for shortened hemodialysis treatment time), automatic ultra filtration control, standard and variable bicarbonate and sodium capabilities and automated chemical disinfection. The Meridian will operate in either the bicarbonate or acetate mode for concentrates. Recall # Z-0605-03.

    This is a Class III recall - a situation in which use of or exposure to a violative product is not likely to cause adverse health consequences.

    CODE
    Product codes 5M5576, 5M5576R,
    serial numbers 200000 through 202864.

    RECALLING FIRM/MANUFACTURER
    Baxter Healthcare, Corp., Round Lake, IL, by letter dated November 15, 2002. Firm initiated recall ongoing.

    REASON
    Meridian hemodialysis instrument may reset when in shutdown mode causing the system to reset with potential for patient increased bioburden exposure.

    VOLUME OF PRODUCT IN COMMERCE
    2,864 units.

    DISTRIBUTION
    Nationwide, Mexico, China, South Korea, Hong Kong.

March 6, 2003 - There is one recent medical product recall of interest to the dialysis community in the FDA Enforcement Report for March 5, 2003:

  • PRODUCTS:

    • HomeChoice Automated Peritoneal Dialysis Systems; Made in U.S.A.; catalog numbers 5C4471, 5C4471R, 5C4474, 5C4474R, T5C4441, T5C4441R. Recall # Z-0564-03;
    • HomeChoice PRO Automated Peritoneal Dialysis Systems; Catalog numbers 5C8310, 5C8310R, R5C8320, R5C8320R, T5C8300, T5C8300R. Recall # Z-0565-03.


    This is a Class III recall - a situation in which use of or exposure to a violative product is not likely to cause adverse health consequences.

    CODE
    a) All HomeChoice Systems with software versions 8.51 and 8.52;
    b) All HomeChoice PRO Systems with software versions 8.51 and 8.52.

    RECALLING FIRM/MANUFACTURER
    Recalling Firm: Baxter Healthcare, Corp., Round Lake, IL, by letters dated December 23, 2002.

    Manufacture: Baxter Healthcare, Corp., Largo, FL. Firm initiated recall is ongoing.

    REASON
    Potential for low ultrafiltrate volumes when using the Low Fill Mode during continuous cycling peritoneal dialysis (CCPD) therapy.

    VOLUME OF PRODUCT IN COMMERCE
    12,667 units.

    DISTRIBUTION
    Nationwide and Internationally.


February 21, 2003 - There are two recent medical product recalls of interest to the dialysis community in the FDA Enforcement Report for February 19, 2003:

  • PRODUCT: Various Baxter solution administration sets

    This is a Class I recall - a situation in which there is a reasonable probability that the use of or exposure to a violative product will cause serious adverse health consequences or death.

    a) Baxter Vented Buretrol Volumetric Pump Solution Set, 10 drops/mL; catalog 1C8384; 10 units per case. Recall # Z-0504-03;

    b) Baxter Interlink System Minivolume Extension Set;catalog 1C8412. Recall # Z-0505-03;

    c) Baxter Y-Type Minivolume Extension Set; catalog 1C8499. Recall # Z-0506-03;

    d) Baxter Interlink System Extension Set with 0.22 Micron Filter, catalog 1C8546. Recall # Z-0507-03;

    e) Baxter Interlink System Vented Buretrol Solution Set, 10 drops/mL; catalog 1C8560. Recall # Z-0508-03;

    f) Baxter Interlink System Minivolume Extension Set with 0.22 Micron Filter, catalog 1C8577. Recall # Z-0509-03;

    g) Baxter Interlink System Buretrol Solution Set, 10 drops/mL; catalog 1C8633. Recall # Z-0510-03;

    h) Baxter Auto Syringe Micro-Volume Extension Set with 0.22 Micron Filter, catalog 1M8516 and 1M8516Y. Recall # Z-0511-03;

    i) Baxter Auto Syringe Micro-Volume Extension Set with 0.22 Micron Filter, catalog 1M8522 and 1M8522Y. Recall # Z-0512-03;

    j) Baxter Auto Syringe Micro-Volume Extension Set with 0.22 Micron Filter, catalog 1M8527 and 1M85827Y. Recall # Z-0513-03;

    k) Baxter Auto Syringe Extension Set with 0.22 Micron Filter, catalog 1M8529. Recall # Z-0514-03;

    l) Baxter Minivolume T-Connector Extension Set with 0.22 Micron High Pressure Pediatric Filter, catalog 2C5683, 48 units per case. Recall # Z-0515-03;

    m) Baxter Minivolume Extension Set with 0.22 Micron High Pressure Pediatric Filter, catalog 2C5693, 48 units per case. Recall # Z-0516-03;

    n) Baxter Interlink System Microbore T-Connector Extension Set with 0.22 Micron High Pressure Pediatric Filter, catalog 2N3332, 200 units per case. Recall # Z-0517-03;

    o) Baxter Interlink System Microbore T-Connector Extension Set with 0.22 Micron High Pressure Pediatric Filter, catalog 2N3333, 200 units per case. Recall # Z-0518-03;

    p) Baxter Microbore 3-Lead Extension Set with 0.22 Micron High Pressure Pediatric Filter, catalog 2N3342, 50 units per case. Recall # Z-0519-03;

    q) Baxter 36" Micro-Volume Extension Set, catalog 2N3347. Recall # Z-0520-03;

    r) Baxter 60" Micro-Volume Extension Set, catalog 2N3350. Recall # Z-0521-03;

    s) Baxter Interlink System Minivolume Extension Set, catalog 1C8483. Recall # Z-0522-03;

    t) Baxter Interlink System Buretrol Solution Set, 60 drops/mL; catalog 1C8600. Recall # Z-0523-03.

    CODE
    All lots received prior to 10/23/2002.

    RECALLING FIRM/MANUFACTURER
    Recalling Firm: Baxter Healthcare Corp., Round Lake, IL., by letter dated October 17, 2002.

    Manufacturer: Baxter Healthcare Corp., Deerfield, IL. Firm initiated recall ongoing.

    REASON
    The I.V. filter sets contain a 1.2 micron filter instead of the labeled 0.22 micron filter.

    VOLUME OF PRODUCT IN COMMERCE
    a) 2,064 units;
    b) 679,968 units;
    c) 13,104 units;
    d) 0 units;
    e) 21,972 units;
    f) 1,920 units;
    g) 4,908 units;
    h) 16,860 units;
    i) 53,310 units;
    j) 0 units;
    k) 0 units;
    l) 198,816 units;
    m) 634,992 units;
    n) 47,000 units;
    o) 265,240 units;
    p) 130,040 units;
    q) 247,800 units;
    r) 630,140 units;
    s) 74,832 units;
    t) 9,072 units.

    DISTRIBUTION
    Nationwide and Internationally.


  • PRODUCT: Lifepak 12 defibrillator/monitor - a portable battery operated, complete acute cardiac care response system with both manual and semi-automatic defibrillation operation. Recall # Z-0542-03.

    This is a Class II recall - a situation in which use of or exposure to a violative product may cause temporary or medically reversible adverse health consequences or where the probability of serious adverse health consequences is remote.

    CODE
    All LIFEPAK 12 units are affected.

    RECALLING FIRM/MANUFACTURER
    Medtronic Physio Control Corp., Redmond, WA, by letter, on January 27, 2003. Firm initiated recall is ongoing.

    REASON
    Impact and damage to the high voltage connectors results in poor or no electrical contact from the cable to the LIFEPAk 12 unit.

    VOLUME OF PRODUCT IN COMMERCE
    46,042.

    DISTRIBUTION
    Nationwide and Internationally.



[This message has been edited by Gary Peterson (edited 02-23-2004).]
 