<Mr. Mech>
Posted
Certain stations run high no matter what machine is put there.
pressure at port seems ok (3-6psi ). Machines run fine on acid jugs. wtf
 
Reply With QuoteReport This Post
<adam>
Posted
Install white check valves on the inlet side of the acid and bicarb pumps. This will decrease the pressure from loop. The head tanks at one of my clinics are almost 20 ft high so there is alot of pressure. Installing the white checks valves brought my conds down from 14.0 to 13.6
 
Reply With QuoteReport This Post
Posted Hide Post
are you using the strong spring in your acid pump?
 
Posts: 192 | Location: Ann Arbor, MI | Registered: 16 September 2009Reply With QuoteReport This Post
<Mr. Mech>
Posted
Machines already have check valves installed, even put in new ones. We have olc/diasafe so the pumps have two weak springs. The problem is its only at two or three stations, no matter what machine is put there.
 
Reply With QuoteReport This Post
<Someguy>
Posted
just to clarify-They run fine on jugs at those stations or in the machine repair room?
Thanks
 
Reply With QuoteReport This Post
<Mr. Mech>
Posted
The problem is on the floor, the machines run at 14.6 from the wall, when they plug the acid in a jug it drops to 14.0 or less, in the tech area everything is on jugs and normal. Have already tried pumps, check valves, calibrations, pressure at wall is 4-6 psi, checked drains----everything seems normal, even replaced drain connection, we have quick disconnects, thought maybe it went bad, no change
 
Reply With QuoteReport This Post
<Someguy>
Posted
I'm stumped. Couldn't be drain restriction if they run fine on jugs at that station. Can't think of anything besides pressure sorry. Confused
 
Reply With QuoteReport This Post
<wall pressure>
Posted
reduce wall pressure see if that helps. use a regulator if you got one crank it down to a drip to rule out pressure
 
Reply With QuoteReport This Post
<oldguy>
Posted
Had the same problem cal. acid and bicarb pumps without check vales in line and then put back in after cal and try on that station work for me
 
Reply With QuoteReport This Post
<Jed>
Posted
Mr. Mech is your acid loop a vertical feed system if it is check the presure coming out of the wall you should get about 4-6 psi If it is realy high you might want to lower it to reduce presure coming out of the wall.
 
Reply With QuoteReport This Post
<breakpar now>
Posted
Mr Mech - FYI - if you have to calibrate your acid pump with the Buret, do it without the check valve. Otherwise you won't get an accurate reading.
 
Reply With QuoteReport This Post