Has anyone ever seen the "INTERRUPTED" message pop up when an "H" is in either the rinse cycle or the heat disinfect cycle. I have one and can't figure out what would cause that.
Thanks in advance for any help.
 
Posts: 14 | Location: Memphis, TN. USA | Registered: 21 December 2003Reply With QuoteEdit or Delete MessageReport This Post
<pato>
Posted
cdrcg,
You may have a defective "Reed Switch Assembly". Also, either of the wands may not stay in place during those programs.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
But it only happens when the shunt door is opened. I forgot to mention that at first.
 
Posts: 14 | Location: Memphis, TN. USA | Registered: 21 December 2003Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
check the 3 micro switches on the shunt door. sometimes they can stick or the springs on the plungers brake. good luck
 
Posts: 55 | Location: Midwest | Registered: 28 February 2005Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Thanks! I knew there would be a simple solution, I just couldn't come up with it. I'm sure it will be the microswitches because the springs still seem strong. Thanks again!
 