What is the temp of the Gambro Phoenix that it reaches during a Thermal??

Does anyone have any documentation on how high the temp of the machine needs to be for a proper disinfection? It would be wet heat

Thank you in advance.
 
Reply With QuoteEdit or Delete MessageReport This Post
<jdbiomed>
Posted
The "benchmark" for Disinfection temp on 200K is 80 degrees C.
 
Reply With QuoteEdit or Delete MessageReport This Post
<North>
Posted
The Phoenix typically runs about 95 degrees C during the heat disinfect cycle.
 
Reply With QuoteEdit or Delete MessageReport This Post