<kennyG>
Posted
Is there a standard on leakage current, grounding, that sort of thing. If so, Where can
I find it?
 
Reply With QuoteEdit or Delete MessageReport This Post
<Atlanta Tech>
Posted
Electrical Leakage Requirements are noted in ANSI/AAMI RD5, under Electrical Safety Requirements. The standard (also noted) is: Complaince with the NON-ISOLATED patient connection requirements of SAFE CURRENT LIMITS FOR ELECTRICAL APPARATUS.
Good luck with all your endeavors
 
Reply With QuoteEdit or Delete MessageReport This Post
<Steve>
Posted
The standard is .5 ohm from power cord ground to device chassis. Leakage current is <300 uA's in patient care areas and < 500 uA's in non-patient care areas. A patient area is defined as within six feet of a patient. (logic being that a doctor or nurse could reach out from a device and touch a patient).
 
Reply With QuoteEdit or Delete MessageReport This Post
<kennyG>
Posted
My humble thanks to all who replied. I deeply appreciate it.
 
Reply With QuoteEdit or Delete MessageReport This Post