I have a flow error on a k machine with alot of hours on it ,the condo only goes to 12.5 and no higher,I have done all the calibrations and changed the acid pump.and check valves.There is'nt a constent error just on and off.....Please help
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
balance chambers!!!
 
Posts: 48 | Registered: 20 February 2006Reply With QuoteEdit or Delete MessageReport This Post
<bamatech>
Posted
Check the flow from your drain. If it has a hard pulse and then a dribble our a couple of pulses and then a dribble you need to change the membrane in your balance chamber. Drain flow should always be steady even pulses of fluid.
 