My company has purchased Aquaboss RO's for use in a new hospital. The label on the last pressure gauge says 20-35 PSI. There is no way I will have that much pressure after two carbon tanks, the narrow tubing Fresenius uses now and two filters.
How much pressure do you really need to operate the Aquaboss?
<Guest>
Posted
Mine runs about 25 to 30 psi and 14 to 15 psi when filling. I do not have any alarms so I guess that works. As long as your break tank fills before it cycles you shouldn't have any problems. If you do, Lauer sells a pump for that.