<DaVita>
Posted
I've got a K with about 3800 hrs. Just changed the diasafe, bleached & rinsed then ran the diasafe test. It failed x 3. Put another new diasafe in & went thru the same process & it keep's failing. Any ideas?
 
Reply With QuoteEdit or Delete MessageReport This Post
JP
Posted Hide Post
I would investigate the possibility of a hydraulic leak elsewhere in the system.
 
Posts: 51 | Registered: 15 January 2008Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
Check for a pinched valve cable at the distribution board or try a new valve.
 
Reply With QuoteEdit or Delete MessageReport This Post
<old F Guy>
Posted
Will the machine pass a pressure test? More than likely you will find it also fails neg pressure, look for leaks in the hydraulic system.
 
Reply With QuoteEdit or Delete MessageReport This Post
<DDC@UCDMC>
Posted
You say Diasafe. If you have Diasafe you will have four o rings and quuick connectors. You may have a leaking o ring??? Or the tubing to the connector may be loose???? Diasafe plus
is without o rings. Look for air in the sight glass on the blue dialysate line. Some tdimes these machines have a way of humbling us.
 
Reply With QuoteEdit or Delete MessageReport This Post
<DDC@UCDMC>
Posted
You say Diasafe. If you have that then you should have Four Quick connectors and o rings??? You may be leaking from an o ring or the tubing to the connector. Look for air in the sight glass on the blue Dialysate Connector.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Razor>
Posted
Check UF circuit for leaks and then calibrate UF volume. I had the same problem last week, turned out to be a leak in the UF circuit.
 
Reply With QuoteEdit or Delete MessageReport This Post
<southerntech>
Posted
Check the Transducer by the diasafe filter. It's probably wet. This will cause failure of the test.
 
Reply With QuoteEdit or Delete MessageReport This Post
orc
Posted Hide Post
As Southerntec says, if the transducer is wet it will fail. If it is wet take it off, shake the water out and re-test. It should pass. The only thing I do not know is why it gets wet in the first place. Anybody know where you can get a copy of an explaination of the test?
 
Posts: 361 | Registered: 20 January 2005Reply With QuoteEdit or Delete MessageReport This Post
<toledo>
Posted
Yes like our guest said The wires going to the valve seem to get pinched if your not careful when puttting the cover on. Check the wires where the cover and the case of the distribution board meet. If not, then the valve, then control and so on. easy fix
 
Reply With QuoteEdit or Delete MessageReport This Post
<XxROOKIExX>
Posted
I too have found failing diasafe test can be a result of pinched wires in the distribution board.
 
Reply With QuoteEdit or Delete MessageReport This Post