I had a alarm that said dial valve failure
1 anyone knows what this is?
 
Reply With QuoteEdit or Delete MessageReport This Post
<Techmex>
Posted
Is there a valve/flow error noted in the status screen?
 
Reply With QuoteEdit or Delete MessageReport This Post
ss
Posted Hide Post
Usually these errors are caused by a sticking valve; or switch in the shunt door. I would check valve 24 and valve 26, then try the switch in the shunt door.
 
Posts: 228 | Registered: 12 November 2008Reply With QuoteEdit or Delete MessageReport This Post
<a42>
Posted
This problem appeared on the screen for a couple of seconds and went away, did not get to see if there was a valve problem or flow problem
 
Reply With QuoteEdit or Delete MessageReport This Post
ss
Posted Hide Post
Just try what I said. If it doesn't come back then don't worry about it. It is a fatal error and you will have to shut off the machine to bypass it to return blood.
 
Posts: 228 | Registered: 12 November 2008Reply With QuoteEdit or Delete MessageReport This Post