<Biomed1313>
posted
Does anyone ran into an issue for no chemical intake? When we hook up the uptake wand, no Minncare get pulled into the RO. Then we get stuck with one hour rinse.
 
Reply With QuoteReport This Post
<emg>
posted
the chemical intake pump is not working properly, membrane inside the mechanism maybe sticking.
1. try to prime the chemical pump by pushing a r.o. water in the intake side using a syringe (apply a bit of pressure to force the membrane to open) during chemical disinfection then replace the tubing back the port and continue the disinfection using minncare.
2. replace chemical pump.
 
Reply With QuoteReport This Post
<Guest>
posted
Does the chemical wand have a O ring on it that may need replaced ?
 
Reply With QuoteReport This Post
<emg>
posted
Yes there should be o-ring otherwise it will not suck the chemical.
 
Reply With QuoteReport This Post
<ship>
posted
I called for the oring and they would not give me the part number for just the oring instead I had to buy the whole wand which is not cheap
 
Reply With QuoteReport This Post
<emg>
posted
just find any o-ring that fits the wand end groove and seals the path.
 
Reply With QuoteReport This Post
<Guessed>
posted
Like an acid oring, or bicarb oring?
 
Reply With QuoteReport This Post
<Guest>
posted
You may need to force chemical into the wand by using a syringe.
 
Reply With QuoteReport This Post
<Old Tech>
posted
WT100319020 is the chemical wand o-ring
 
Reply With QuoteReport This Post