I have a 2008k machine the blood leak and dimmness will let me cal. it and work fine for about a week then come back un calibrated. what would cause that to happen ?
 
Posts: 24 | Location: Houston, Texas | Registered: 18 September 2007Reply With QuoteReport This Post
<trw>
Posted
had one do this a few months ago ended up being the functional board. Wouldnt hold cal
 
Reply With QuoteReport This Post
Posted Hide Post
Yup i had the same issue, (Functional Board)


"Fresenius" the Dialysis God
 
Posts: 5 | Location: Los Ange-less | Registered: 19 March 2008Reply With QuoteReport This Post
<1234>
Posted
Had this on almost all of my new K machines. I took the blood leak detector out and found finger prints on almost all of them.
 
Reply With QuoteReport This Post
JP
Posted Hide Post
I have had two of these and the funky board always fixed it.
 
Posts: 51 | Registered: 15 January 2008Reply With QuoteReport This Post
<jjrunning>
Posted
Had the same issue as 1234. Drove me nuts until the dang things were cleaned.
 
Reply With QuoteReport This Post