Guest can i get a copy too? jazzescoto@hotmail.com Thanks!
|
||
|
| <Mark Halloran>
|
North,
New Interpretive guidelines, V195: "...each adsorption bed shall have an EBCT of at least 5 minutes at the maximum product flow rate (a total EBCT of at least 10 minutes". My understanding is that tanks in parallel are considered to be a "bed", each tank in a series configuaration is also a "bed". If you need three of your four "in series" tanks to get 10 minutes, then one or more of them is less than 5 minutes and therefore not in compliance. |
||
|
MarCor Purification 1-800-346-0365
Hope this helps ya ... 335-10-013 pump decal set $9.00 335-07-121 directional flow arrows $0.14 335-07-123 pump on decal for storage tank$0.64 335-07-124 pump off decal for storage tank$0.64 335-07-125 RO on decal for storage tank $0.64 335-07-126 RO off decal for storage tank$0.64 335-56-512 Set of PG1-PG15 Pressure Gauge ID tags $36.00 335-57-512 Set of TP1-TP12 Test Port ID tags$51.75 335-55-5501 thru 50 MarCor valve tag kit$172.50 335-10-001 sediment filter decal $9.00 335-10-002 water decal $9.00 335-10-003 carbon decal $9.00 335-10-004 multimedia decal $9.00 335-10-005 RO decal $9.00 335-10-006 concentrate decal $9.00 335-10-007 bicarb mixer/storage decal $9.00 335-10-010 brine tank decal $9.00 335-10-011 decal set-RO $9.00 |
||||
|
| <LabelKits>
|
Ameriwater has waterproof label kits for CMS Conditions of Coverage. All components plus directional arrows for pipe flow. 800-535-5585.
|
||
|
| <North>
|
Mark
Please do not make assumptions and then accuse me of not meeting standards. I can assure you we meet EBCT for each tank. The key phrase in the guidelines is "at least". I was just giving an example of how that by exceeding the minimum EBCT it makes it easier to do maintenance during operating hours. |
||
|
North,
Either you mis-stated your post or, Mark is correct. The guidelines state... ...each adsorption bed shall have an EBCT of at least 5 minutes at the maximum product flow rate (a total EBCT of at least 10 minutes". In your post you said... "We only need three to meet EBCT." If you need 3 tanks to meet the EBCT then each tank is giving you less than the required 5 minutes. While you may be meeting the TOTAL EBCT, you are not meeting it for the individual tanks. Chuck DISCLAIMER : My opinions and views are mine and may not be the same as my employer. |
||||
|
| <Mark Halloran>
|
North,
No accusation or offense was intended, it was an observation that your numbers, in a sense, did not add up. V192 is one of the few changes that are clear and easy to read. I consider this forum as kind of an on-going internal audit and I have actually made changes to my policies and procedures based on what I read here. Better we should find something than a state surveyor. |
||
|
One other thought occurred to me regarding the carbon tank configuration NOrth describes at his 45 station unit. If every individual carbon tank can be bypassed, does that mean all of the carbon tanks could be bypassed, inadvertently? Aren't the carbon tanks supposed to be plumbed to make this impossible?
|
||||
|
| <Mt Goat>
|
I used the Ameriwater labels. They look good.
|
||
|
Mt Goat,
Agreed, the AmeriWater labels that came with the last system I bought from them 5 months ago "looked" great. Unless they have changed since then, they do not meet the criteria of the new regulations. Chuck DISCLAIMER : My opinions and views are mine and may not be the same as my employer. |
||||
|
| <labels>
|
The new AmeriWater labels (just released within the past two months) are designed specifically for the new requirements. Previous labels were not designed for this purpose.
|
||
|
| <Guest>
|
Can you have signs made up and posted on the wall behind each piece of equipment or do they have to be plastered on the equipment
|
||
|
"Guest"
IMHO they should be on the equipmentbut in actuallity it probably dosn't really matter, it's all going to depend on the surveyors individual interpretation. AAMI RD52 states... "If water system manufacturers have not done so, users should label major water system components in a manner that not only identifies a device but also describes its function, how performance is verified, and what actions to take in the event performance is not within an acceptable range." Chuck DISCLAIMER : My opinions and views are mine and may not be the same as my employer. |
||||
|
| <Guest>
|
How are they hooked up? What's your Empty bed contact time? Worker and Polisher are to be of equal size with a 10+ minute EBCT spread over the two stages. Could be Worker, Polisher, and Spare maybe.
|
||
|