Technical gurus,

After an RO interlock correction, seven of fourteen H machines (supposedly?) alarming "Inlet flow Error". However, after an hour, they correct themselves. Has anyone ever seen this? For the record, what are the first possible problems and corrective measures? Lastly, what part/parts would you suspect and have on hand for corrective measures?

Thank you for your in-put
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
This usually happens when the dearation pump gets air locked after "no water" alarms for too long. Once the water supply is restored after the extended "no water" event, the hydroblock is full, but the dearation pump isn't taking any because it's air locked. If it's not taking any water from the hydroblock, the float switch stays up. If the float switch stays up, you get the "flow inlet error". Happens a lot in acutes when their portables shut down due to low incoming water pressure.

As for a fix, the only real way to "fix" it is to wait and see if the air lock clears or to take the back panel off, get the air out of the dearation pump inlet tubing, reconnect it, disinfect your machine, and set it back up. There's no real "easy" way to fix it but to try to keep it from happening in the first place.

Hope that helps.


Some people dream of success, while others live to crush those dreams.
 
Posts: 575 | Location: Midwest | Registered: 22 December 2002Reply With QuoteEdit or Delete MessageReport This Post
<Atlanta Tech>
Posted
Thanks GTSCSS
 
Reply With QuoteEdit or Delete MessageReport This Post
 Previous Topic | Next Topic powered by eve community  
 

RenalWEB Home Page    renalweb.groupee.net    RenalWEB Discussion Forums  Hop To Forum Categories  Dialysis Technician Forums  Hop To Forums  Dialysis Technical Forum    Pardon the Error, "Inlet Flow Error" was the alarm.

Copyright RenalWEB 2008