<Mr. Mech>
Posted
Have a Fresenius 2008K w/ internal diasfe filter, has about 20,000 hrs, failing diasfe test, replaced filter and transducer. Positive pres test revealed a few small leaks, still fails. Runs fine in dialysis mode and passes all tests. Did notice a flow error when put into diasfe test, not sure if thats normal.
 
Reply With QuoteEdit or Delete MessageReport This Post
<ohmygolly>
Posted
check you deair and flow pressures and valves
 
Reply With QuoteEdit or Delete MessageReport This Post
<Mr. Mech>
Posted
Checked pressures, all were OK, whats next
 
Reply With QuoteEdit or Delete MessageReport This Post
<Mr. G>
Posted
24 psi on the loading pressure....
 
Reply With QuoteEdit or Delete MessageReport This Post
<breakpar now>
Posted
Check to see if your distribution board cover has pinched the wire going to the valve on the diasafe.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Mr. Mech>
Posted
It has the manual valve, not the solenoid
 
Reply With QuoteEdit or Delete MessageReport This Post
<Down Soth Tech>
Posted
I have seen that tie wrapping the tubing on top of both condo cells and top of air seperator will stop these failures.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Mr. Mech>
Posted
Calibrated the dialysate pressure and it passed 3x. You shouldn't have to put wire ties on anything that didn't come that way, no offense but just cut the tubing back or replace
 
Reply With QuoteEdit or Delete MessageReport This Post