I have a K mahine that will go through the entire heat disinfect and with 1 second left on the timer give me a audible alarm with no errors and freeze. I think I remember someone posting about this problem some time ago but I couldn't locate it in the archives.
 
Posts: 182 | Location: Arkansas, USA | Registered: 14 January 2003Reply With QuoteReport This Post
<nonameuno>
Posted
I have had a bad power logic board cause this.
 
Reply With QuoteReport This Post
Posted Hide Post
quote:
<nonameuno>


Thanks,
I'll give that a try.
 
Posts: 182 | Location: Arkansas, USA | Registered: 14 January 2003Reply With QuoteReport This Post
<southerntech>
Posted
Just reseat the power logic board. Have seen this several times and reseating works.
 
Reply With QuoteReport This Post
Posted Hide Post
I changed out the PLB. I had previously reset it and it didn't get the kink out. Thanks Guys. Another problem resolved thanks to you and the community!
 
Posts: 182 | Location: Arkansas, USA | Registered: 14 January 2003Reply With QuoteReport This Post
<IC 9>
Posted
If IC 9 is socketed on the old power logic board try replacing it. On newer K machines it may be soldered directly on without a socket. Also, you can use an H power logic board in a K machine, but you cannot use a K power logic board in an H. The H board has an extra 12 volt circuit on it that is not needed in a K machine. That's why Fresenius is charging $314 for the H board and $537 for the K board, even though the K board has less on it. Go figure, I guess that's why my Fresenius stock keeps going up!
 
Reply With QuoteReport This Post