Need some advice on a high LAL and colony count on a GE portable R.O.. Here is where I am at first test result was Colony Count of > 600 CFUs 5.640 EU/mL drawn on 6/11. I disinfected with Minncare for 1 hour and repeated. Next result was 94 CFUs and 0.458 EU/mL drawn on 6/15. I disinfected with Minncare for 5 hours and repeated. Checked results today on website for lab so far the LAL is 0.108 EU/mL have not yet recieved Colony Count this was drawn on 6/22. Any advice or suggestions would be greatly appreciated.
<bikertech>
Posted
How old are the membranes?? It may be time to replace them.
<Guest>
Posted
Need more info. how old is portable or when membrane last changed? Did you hi an lo pH clean before disinfect?
How is the product water conductivity/TDS and % rejection?
If the water quality hasn't decreased I would think your membranes are still good but became colonized. Your lab results so far look as if your disinfections are working.
Chuck
DISCLAIMER : My opinions and views are mine and may not be the same as my employer.
Posts: 1095 | Location: Baltimore, MD USA | Registered: 24 October 2001
My machine is 28 months old with 2021 hours on it. this is in an acute unit so is sometimes idle for long periods of time. I did do a high and low PH cleaning prior to disinfect. It does run once a day for 1/2 hour to disinfect the dialysis machine that is hooked up to it. Thanks everyone for the fast responses.