<nerf>
Posted
Morning All,
Could use some help its been a while since I had to fool with this but I have a H machine that will not hold the concentrates I calibrate them in then once I go back into regular mode it tells me to enter concetrates if Iam not mistaken this information is stored on the battery ram on the function board right?
Thanks for any help
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
When you are through entering concentrates are you turning the machine off,flipping the switch to go back to dialysis or are you backing all the way out to the initial service mode screen before turning it off? If you just turn it off you are not saving the data,you have to back all the way to the initial service mode screen.
 
Posts: 85 | Location: TEXAS | Registered: 23 January 2006Reply With QuoteEdit or Delete MessageReport This Post
<nerf>
Posted
I am backing out by hitting the reset button to go back to the calibtrate start screen
 
Reply With QuoteEdit or Delete MessageReport This Post
<Atlanta Tech>
Posted
nerf,

Sounds like your machine is losing it's mind. The function Bd. EPROM holds most of the calibrations. You can try this to check out weather it's memory, or the board itself. Swap that EPROM to a known working function Bd. then try saving concentrate info. If the memory holds, bad board, If lost, bad EPROM. When backing out of the enter concentrate field, make sure your choices are highlighted, before you start pressing reset. Let us know what you find. Also, if it's a board you need Fresenius is asking $900 bucks for those, I hope you can utilize another vendor (H&S Technical, perhaps).
 
Reply With QuoteEdit or Delete MessageReport This Post
<g>
Posted
WRONG!
Eprom on H functional bd holds no calibrations.
BatRam IC 5 hold calibrations.
 
Reply With QuoteEdit or Delete MessageReport This Post
<botech>
Posted
Before you start messing around with the Battery Ram ( <g> is correct about the cal.'s),I believe you not only have to back out to the calibration/ diagnostics/options screen, but turn the machine off, and then restart for the concentrate info to be saved.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
IC5 is a NVRAM which stands for Non Volatile Random Access Memory. It requires no external power to store data. It holds 32K x 8 bits of memory.
 
Posts: 48 | Location: Hawaii | Registered: 12 September 2007Reply With QuoteEdit or Delete MessageReport This Post
<F-tech.>
Posted
Just asking, Are you entering Concentrates? If I remember correctly with our Hs is we enter and then highlight the concentrates used, once highlighted it should apear in the dialysis concentrate verification screen in Dialysis mode.Its just a check I do before going into calibrations.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
you need to press the set button on the machine till you return to the main menu. if you reset it may not hold the calibration. Also press the set button once then once then once till you exit. dont press them fast.
 
Posts: 7 | Location: I-R-AZ | Registered: 06 May 2008Reply With QuoteEdit or Delete MessageReport This Post
<Medtech1437>
Posted
Sounds like the bat ram is faulty. I would try swapping a known good Function Board in TNC mode of course and see if it holds calibration.
 
Reply With QuoteEdit or Delete MessageReport This Post
<ladytech>
Posted
WinkSounds like the battery ram may be full,years ago we had to clear out the battery ram to allow us to save. I do not remember of course the procedure for this, Where oh Where are you KTECH I am sure that you can jog my memory Roll Eyes
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Not much help here on this, because we have not entered more than 4 or 5 concentrates on any of our machines. But you can certainly delete concentrates that you are not using and see if that helps with the problem. I don't know the maximum number that can be entered. I suspect that the problem originally was operator error, like not hitting SET at the right time, as I have had the same thing happen to me when entering concentrates.
 
Posts: 367 | Location: Cleveland.OH | Registered: 03 February 2006Reply With QuoteEdit or Delete MessageReport This Post