RenalWEB Home Page    renalweb.groupee.net    RenalWEB Discussion Forums  Hop To Forum Categories  Dialysis Technician Forums  Hop To Forums  Dialysis Technical Forum    fresenius K actuator temp low failure

Moderators: Dennis Todaro
Go
New
Find
Notify
Tools
Reply
  
-star Rating Rate It!  Login/Join 
Posted
I have a K machine failing actuator temp low. The power logic bd was replaced by another tech. I have since changed ntc 2&3 (& cal'd them), sensor bd and cable,act. bd, verified pressures, heater rod, with no resolution- anyone have any other ideas?


Say what you'll do & Do what you say.
 
Posts: 3 | Registered: 06 August 2008Reply With QuoteEdit or Delete MessageReport This Post
JP
Posted Hide Post
Is this one of those that has a combination power logic and test board? If not the case I would change the test board. Were all of these parts known good or “new” from FMC? Other than that, I am not sure.
 
Posts: 45 | Registered: 15 January 2008Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
quote:
Originally posted by JP:
Is this one of those that has a combination power logic and test board? If not the case I would change the test board. Were all of these parts known good or “new” from FMC? Other than that, I am not sure.
That new board is a combination actuator and test board, not the power logic and test board. So have you tried a test board? That K test board is the same as an H machine if that helps.
 
Posts: 367 | Location: Cleveland.OH | Registered: 03 February 2006Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I tried the test board with no luck. A bit of creative troubleshooting showed the functional board to be the culprit. These K's are actually 2kos's so the boards hadn't been combined yetSmiler


Say what you'll do & Do what you say.
 
Posts: 3 | Registered: 06 August 2008Reply With QuoteEdit or Delete MessageReport This Post