I have 2 K2 units only 2 1/2 mo. old showing "Rinse high cond" alarm. The problem is it only shows when I power up. Go to Rinse or Dialysis the alarm goes away. Anybody seen this before or have come across this problem?
 
Reply With QuoteEdit or Delete MessageReport This Post
<John>
Posted
F-Tech,

This could be a result of the machine sensing a chemical, commonly bleach, residually left in the machine. If you have bleached the system, and rinsed the system, the machines would show this when you power it on. Check for the presence of bleach at the machine drain or sample port for less than 0.5 ppm. If so, press "Reset", and the alarm should go away.

- John
 
Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
I get this at start up myself on occasion. 100% of the time for me, the nurses did not have the red or blue connector fully inserted during rinse. I push it back in, hit reset and rerinse.
 
Reply With QuoteEdit or Delete MessageReport This Post
<F-tech.>
Posted
Thanks guys, I checked for bleach and pushed in the red connectors ( wich was not fully inserted ) reset and rinse, problem gone. I was about to call Fresenius and do a shakedown on their tech. support dept. to retrieve the secret into solving this mystery, but no need now we have experts here on the forum.
 
Reply With QuoteEdit or Delete MessageReport This Post
<tech in ilinois>
Posted
I had the same problem. Check the functional board.
 
Reply With QuoteEdit or Delete MessageReport This Post
<BayAreaBio>
Posted
Make sure the techs all clicking the wands all the way into the front ports or else the wont rinse the lines out
 
Reply With QuoteEdit or Delete MessageReport This Post
<Glenfrommen>
Posted
Who said Check the F-Board? Seriously? You hit reset...omg
 
Reply With QuoteEdit or Delete MessageReport This Post
<Guest BIO>
Posted
When u first turn mach on and it shows up push the reset burtton and see if it goes away. Had the same problem on an H mach everytime i cut it on it said that. I push the reset button never came back.
 