while calibrating dialysate pressure for a TMP problem, when u try to get 0 0mmHG on the external pressure gauge the box should read approx 90-100 mine reads about 120 and for -250 it should read approx 140 and mine is 160 i have calibrated venous and then dialysate for high tmp and i cant figure it out it was passing test before calibrating and now its failing actuator high tmp any suggestions asap lol
GOOD MORNING, I would also try transducer #9. KEEP US INFORMED
<Biomed Dude>
Posted
When I calibrate the ZERO point for TMP, I just open it to atmosphere. I don't pull or push to make it read zero. I have never had an issue with TMP... that wasn't PCT induced. Another way to check your zero point for TMP is with it up and running, put the Hansons in a bucket of water and the TMP should be around 40 to 50. The old E's you wanted it to be around 70.
Just had this problem on 2 machines.One was a sensor board the other was the Function board. I would start with the Function board its cheaper.You are right on the values that you should see.
Posts: 12 | Location: DOWN SOUTH | Registered: 05 May 2005