Posted
hi everyone i have a h machine thats showing high and low flow error, my temp is 33.0 and my cond. is 17.0.
any idea thank you
 
Reply With QuoteEdit or Delete MessageReport This Post
JP
Posted Hide Post
I would check the balance chamber membranes.
 
Posts: 45 | Registered: 15 January 2008Reply With QuoteEdit or Delete MessageReport This Post
<botech>
Posted
Could be a couple of things. First check your flow, degas and loading pressures. Check your CFS numbers ( debug screen 10), the switch may not be working. One of your balance chamber valves may not be working correctly. Watch debug screen 20 and see if cycle 1 or 2 on the balance chamber valves is when you get your flow messages. You may also have a torn BC diaphragm. Run a BC volume test. Those are the areas I check.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
definitely check your balancing chamber diaphrams, temp not rising above 33 and cond. at 17, fresh and spent sides are probably exchanging..
 
Posts: 24 | Registered: 08 June 2008Reply With QuoteEdit or Delete MessageReport This Post
<lynn>
Posted
was the machine on the floor. or in the bio room? make sure the machine was not on bypass for a long time.
 
Reply With QuoteEdit or Delete MessageReport This Post
orc
Posted Hide Post
Balancing chambers are not switching. If this happens heater is turned off but acid and bicarb pumps continue. BTW, what colour is your Honda?
 
Posts: 303 | Registered: 20 January 2005Reply With QuoteEdit or Delete MessageReport This Post