To all the F2008h gurus I am asking for your help. I have an "H" machine that the level detector will not pass test. I have switched level detectors, changed actuator, mother and test boards. When the machine is in test mode and is testing the level detector it fails to clamp the venous line or shut down the blood pump. It is also failing optical detect and blood leak 1 alarms. I have my theories but I would like to hear yours. Thank you in advance for your help.
 
Posts: 127 | Location: Texas | Registered: 24 January 2001Reply With QuoteEdit or Delete MessageReport This Post
<one tech>
Posted
Just a thought I had the same problem on the optical. And the fix was to readjust the artrial pressure. just a thought
 
Reply With QuoteEdit or Delete MessageReport This Post
<lurker>
Posted
Try putting a known working blood pump module
from another machine, or if you have one in back-up try it.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Thanks, but I already tried that. The module itself is good.
 
Posts: 127 | Location: Texas | Registered: 24 January 2001Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Have you tried the Actuator, Function, and Level Detector ribbon cables?
 
Posts: 875 | Location: Baltimore, MD USA | Registered: 24 October 2001Reply With QuoteEdit or Delete MessageReport This Post
<CP>
Posted
De air brushes can cause alot of problems.
Check the hours on the brushes I replace them after 6000hrs.
 
Reply With QuoteEdit or Delete MessageReport This Post