can anyone tell me what to do with a "act temp hi" failure on a 2008H machine? i've changed the temp sensor probe, calibrated temp sensors, did the dac value, replaced the actuator board... and nothing! can someone help a brother out? thanks!!
<trw>
Posted
If temp and conductivity have been calibrated and are all ok. try test board sometimes it will cause this type of problem. good luck
<botech>
Posted
No other failures during self test? I would agree with trw, put in a known good test board. If that solves the problem, put the original actuator board back in and run self-test.
<Frustrated>
Posted
thanks you 2, i'll give it a try!!
<dibron>
Posted
Also verify the flow pump pressure at 800 ml flow is within limits.I had the same problem a few years ago and eventually that was a bad flow pump head.In service mode was OK(30psi),but in dialysis mode at 800 < 20 psi.
<Robert>
Posted
That is somthing that comes with changing the Funktion board all the time. Put the resisters on it and recal the heat resister lvls.