<Lumpy>
Posted
Hello Everyone :

I have an 2008K with an on going low flow error. The balancing chamber diaphragms were replaced 11-5-08. Occasionally the machine makes a sound like there is a restriction somewhere. I have tried valves 35,37,36 and 38,
CFS, and a different flow motor assembly. The machine will run several hours without a problem, then the low flow error will manifest itself. The problem is intermittent not constant. Can anyone think of something I have not tried?

Thanks,
Lumpy
 
Reply With QuoteEdit or Delete MessageReport This Post
<Razor>
Posted
I've had the same problem, many times and it always turned out to be a balance chamber valve.
Makes a high pitched squeek, then a few seconds later "Low flow" and then goes high cond.
You already changed 4 valves, but it could still be one of the ones you didnt change. 31-32-33-34.
Change one at a time out,see how it goes or go ahead and change them all out and be done with it.
Hope this helps.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
Have you tried doing a manual valve leak test? This is a good way to isolate the bad valve. Make sure and check the balance chamber volume. I realize you just changed them in November but you may have gotten a bad one or something. Worth checking.
 
Reply With QuoteEdit or Delete MessageReport This Post
<breakpar-now>
Posted
Was the flow motor that you put in new? If not it could be that your brushes are getting worn.
 
Reply With QuoteEdit or Delete MessageReport This Post
<je>
Posted
had the same problem it turned out to be the harness on the one of my balancin chambers valves. i troubleshooted by diconnecting each valve one-by-one until the squeakin noise stopped.
 
Reply With QuoteEdit or Delete MessageReport This Post