<Carlo>
Posted
Having a problem with a FMC K machine(21548 hrs). When I go into calib. mode and try to calib. condo cells I get the message "Conductivity Offset Error". I tried swapping another sensor board and still get the same message. Has anyone seen this before?
 
Reply With QuoteEdit or Delete MessageReport This Post
<breakpar now>
Posted
I'm not sure I'm remembering this right, but I think you have to trick the machine by putting a good value ie: 13.5 then hit confirm and go back and do the condo cal again.
 
Reply With QuoteEdit or Delete MessageReport This Post
<conductivity>
Posted
It is your function board!
 
Reply With QuoteEdit or Delete MessageReport This Post
<quest>
Posted
Try deleting the calibration first. (I forget how but someone can post it) This works 25% of the time.
 