2008H machine in heat and i think toward the end machine, what i thought, was shutting off kept beeping and then smelled hot. now i have a constint flow error and tmp pegged at 520 cond.17.0 temp 33. anything would help thanks in advance


One Love, One God and Only One Way
 
Posts: 219 | Location: Indiana | Registered: 25 August 2005Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Do you have flow into the machine? If so, how far have you traced it up to?


I didn't touch anything....
 
Posts: 7 | Location: North East | Registered: 21 September 2006Reply With QuoteEdit or Delete MessageReport This Post
PB
Posted Hide Post
ok now i have no flow error temp is perfect cond is 10.0 my blood pump is not resetting unless i turn the dial to 000


One Love, One God and Only One Way
 
Posts: 219 | Location: Indiana | Registered: 25 August 2005Reply With QuoteEdit or Delete MessageReport This Post
orc
Posted Hide Post
Check all your brds. for any sign of something burnt. More than likely you will find a burnt component on one of them.(ie actuator brd.)
 
Posts: 303 | Registered: 20 January 2005Reply With QuoteEdit or Delete MessageReport This Post
PB
Posted Hide Post
ok i let this machine sit over the weekend and now i have no flow err. temp and cond are normal now just the problem with the blood pump, will not reset unless i turn dial off to 000 then back on runs fine Confused running through test now