I am stuck!we recently upgraded all of our 2008H machines with digital heparin pumps. I have 5 machines that the pumps WILL NOT work in. I have checked the cables,and pumps. I have upgraded all the software in the machines. (ie Functional boards and actuator boards)I have swapped pumps good with bad with no luck. Can anybody think of anything else to try? Thanks in advance. Matt
 
Posts: 124 | Location: ardmore,ok | Registered: 08 December 2002Reply With QuoteEdit or Delete MessageReport This Post
<Bobby B>
Posted
Have you contacted Fresenius to help out? If so what did they say? Sounds like you did everything to me?
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
yes I have contacted Fresenius done everything they said to do with no results.
 
Posts: 124 | Location: ardmore,ok | Registered: 08 December 2002Reply With QuoteEdit or Delete MessageReport This Post
<Bobby B>
Posted
I would call your Freseniu Sales Rep and get them involved..The Sales Rep should be able to get someone to help you or at least get someone to come out and visit...If that doesn't work, keep working up the food chain...It should only take one call...Their service is questionable at best IMHO!!! There is alot of other vendors at there and I know we will be asking them to come in when where ready...Good Luck
 
Reply With QuoteEdit or Delete MessageReport This Post
<K man>
Posted
Make sure the hep pump is properly grounded.
 
Reply With QuoteEdit or Delete MessageReport This Post
<renalman>
Posted
If you havent' already done so, go into Service Mode, then Treatment Options(or possibly Hardware Options). Verify that a Heparin Pump is slected. If it is, make sure that your dip switches on the Fucntional Board are properly set.

Good Luck!!

RenalMan
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
renal man,
The dip switches are set properly and the heparin pump is selected in the service mode.What about the sensor board? Could this be the problem?
 
Posts: 124 | Location: ardmore,ok | Registered: 08 December 2002Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Check the actuator board. I had a machine that the heparin pump would not work on. Some one had installed the wrong actuator board in it. The old style actuator BD does not have the driver chips installed on it. Compare the actuator boards and see if any chips are missing on the ones that don�t work. mybe the sent your the wrong ones
 
Posts: 55 | Location: Midwest | Registered: 28 February 2005Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
bmedtom,
THANKS A MILLION!!!!!That was the problem. I don't know why Fresenius didn't mention that to me. I have only been fighting with this problem about 5 or 6 months! Thanks Again Matt
 
Posts: 124 | Location: ardmore,ok | Registered: 08 December 2002Reply With QuoteEdit or Delete MessageReport This Post