skiwolf,
My FMC sales rep happened to stop by yesterday and we discussed the Diasafe in general then this topic specifically. He said that "they reccomend they get changed out quarterly which would be after 13 bleachings". Check with your sales rep and see if they have any documentation on this. I still think that since bleach is known to strip polysulfone, you may be on the right track. |
||||
|
| <concern>
|
You might want to ask the people in who put your drains in what you can do about the problem. |
||
|
Here is some documentation that we have on the Diasafe filters.
We are using the Diasafe Plus. We received a package insert with the replacement diasafes. This insert was produced by Fresenius Medical Care AG. (Germany) Here are some of the recommendations on the insert. Disinfection according to the operating instructions of the machine is essential. The mandatory filter integrity test must be manually performed on 2008 series machines according to the operating instructions (of the machine). Diasafe must be exchanged after 12 weeks or 100 treatments or earlier if integrity test fails. Disinfection is mandatory after installation. The Diasafe filter Operators instructions for the 2008H & 2008K were included with our machines on delivery. These instructions were produced by Fresenius USA. (P/N 490039 Rev. B) Here are some of the recommendations in this document. Machine disinfection procedure is unchanged. Replace Diasafe every 90 days (3months). Always perform automated or manual pressure holding test before performing filter integrity test. Do a filter integrity test after replacing the filter. Do a filter integrity test every other week. Instructions for the filter integrity test are also included. Implications: In our unit we do 18 treatments per week on each machine (3 times per day x 6 days). The 100 treatment maximum thus works out to about 5 � weeks instead of 12 if the machine is kept in service the whole time. The bi-weekly filter integrity testing represents a considerable investment of time and record keeping. It is possible to do the tests between treatments and thus avoid some of the time penalty of bringing a machine into conductivity and automated pressure holding test. |
||||
|
| <Great White North>
|
Your biofilm drain problem is directly a result to the new bleach cycle incorperated in the 2008K. 60ml of bleach is recirculated and then expelled once at the end of disinfection. Fresenius USA is currently working to change the bleach cycle back to drawing up 200ml and passing it once thru the machine, then draw up 60ml of bleach and recirculate within the machine. Check with Fresenius Engineers in Walnut Creek,CA they have a whole list of problems that we incountered with the 2008K and are now working to solve.
|
||
|
| <Northern Tech>
|
Up here in Canada Fresenius told us that the Diasafe was good for 12 weeks only if you only Bleach once a week as it has to be replaced after 13 bleach cycles. Why I don't know but that is what they told us, which was a problem for us a we bleach in between patient runs so we would replace the filters weekly, large cost. Also as GWN said the K machine doesn't put enough bleach through the system to help with drain problems, sorry but I don't really have any new suggestions for you to try.
|
||
|