Moderators: Dennis Todaro
Go
New
Find
Notify
Tools
Reply
  
-star Rating Rate It!  Login/Join 
<awh3c>
Posted
If there was a suspected pyrogenic reaction, would you return the patient's blood?
 
Reply With QuoteEdit or Delete MessageReport This Post
<Pyrogen Boy>
Posted
The concept is that by returning the blood you are potentially putting MORE pyrogens into the patient, perhaps escalating their physiological response to the pyrogens. We never return the blood.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Atlanta Tech>
Posted
Check your protocol, same non-return policy for bllodleaks. RD-5 or 8 should give you insight as to AAMI recommendations.

I hope this assists you.
 
Reply With QuoteEdit or Delete MessageReport This Post
 Previous Topic | Next Topic powered by eve community  
 