<adam>
Posted
Fresenius K machine giving me an E.08 alarm every morning when the machine is started up and the prime button will not work. I have replaced the LP955 board and still will not go away. But if I turn the machine off then back on this will go away?!?!?,, anyone have any suggestions. the bulletin said it's caused by a 24V problem somewhere i think it says.
 
Reply With QuoteEdit or Delete MessageReport This Post
<mburgess>
Posted
check the ribbon cable to the blood pump module. Also check your 24 volts must be between 22.8 to 28.0 volts.
 
Reply With QuoteEdit or Delete MessageReport This Post
<adam>
Posted
and what if the 24 volts is out of range?
 
Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
if its not one of the board on the blood pump mol. it might be the blood pump it self.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Robb>
Posted
My supervisor replaces the blood pump mod. everytime we get that error. It fixes the problem everytime
 
Reply With QuoteEdit or Delete MessageReport This Post