I have a machine that has a constant 'Flow Error'. Not low flow error, not high flow error, not flow recirc error. Just plain old Flow Error. I checked debug screen 21 (I think that's it) and found V31 and V33 are open constantly. I checked the voltages at the distribution board and both were a constant 24V (give or take approx 0.5V). Actuator board was changed and its still doing it. I'm not quite sure what to make of it because I really haven't had a whole lot of problems like this. If anyone can help let me know. Oh yeah... all pressures were good.
|
|||
|
check both deareation pump and flow pump pressure first. Then if these are ok then go onto trouble shooting the valves. Most likely its the deareation pump. good luck.
|
||||
|
Oops didn�t read that last sentence. Sorry
|
||||
|
Have you followed the trouble shooting guide and if so what did it do/say? I will do some checking and get back with ya
Good Luck Matt |
||||
|
Looks like I missed a valve... turns out it was V39 causing all the comotion. Funny though, in the past V39 has caused a 'Recirc Flow Error' and not just a generic 'Flow Error'. Ya learn something new every day! Thanks anyways Matt and Tom. I appreciate the help.
Matt [This message has been edited by GTSCSS (edited 05-05-2003).] |
||||
|