<rob>
Posted
why does neg pressure pass and then when pos pressure is testing it goes to fail, don't know if this will tell me what my problem is i have tested about 5 times and 2 have passed and then it fails again
 
Reply With QuoteEdit or Delete MessageReport This Post
<rob>
Posted
when it fails it fails neg stabilizer
 
Reply With QuoteEdit or Delete MessageReport This Post
<Guest>
Posted
you have a leak, either fluid or air, in your hydraulics, or possibly the o-rings for your hansen connectors. do an induced pressure test (procedure is in the troubleshooting guide) and look for leaks in the hydraulics and at the hansens. Make sure you give it time like the guide says, or you may miss the leak.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Rob, you are telling us two different failures. Which is it? Failing Negative Stabilization is one, and Positive Flow On is another. And it probably is a small leak somewhere. Also tell us what model of machine and its age or hours. An older H machine could have a small leak on the air seperator, or bad deaereation, maybe even a bad hose T connection. If it's a newer K machine, then look for a bad Hansen O ring, a loose clamp by the flow bob or something else that might be simple. Maybe one reason that you are not getting many replies here is because you haven't given us the basic information that we all need to help diagnose this.
 
Posts: 311 | Location: Cleveland.OH | Registered: 03 February 2006Reply With QuoteEdit or Delete MessageReport This Post
<TechManager>
Posted
Have had two of these recently, and both have turned out to be valve 24 had cracked. Thse are old machines (30,000 hrs) and have the old amber colored valves. look carefully at the stem
at valve 24. These leaks are easy to miss. Good luck!!
 
Reply With QuoteEdit or Delete MessageReport This Post
<old F Guy>
Posted
quote:
Originally posted by Ktech:
Rob, you are telling us two different failures. Which is it? Failing Negative Stabilization is one, and Positive Flow On is another. And it probably is a small leak somewhere. Also tell us what model of machine and its age or hours. An older H machine could have a small leak on the air seperator, or bad deaereation, maybe even a bad hose T connection. If it's a newer K machine, then look for a bad Hansen O ring, a loose clamp by the flow bob or something else that might be simple. Maybe one reason that you are not getting many replies here is because you haven't given us the basic information that we all need to help diagnose this.


I think what Rob is talking about the machine shows it has passed neg pressure test on the test screen but then goes back and changes it to failed while the machine is in the pos press test. I've seen this several times. You've got to be watching the test screen at the right time to see it. Like the others have posted you have a leak somewhere. The passed switching to failed thing is not out of the norm with many K machines. Many just haven't watched the screen that close during test I guess.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Yes I have seen the pass and then fail occur. Do you have the diasafe filter on this machine. Start there even if it is a newly changed filter. There have been times when I just repositioned the filter and it passed. I changed the diasafe filter and all was good.
 
Posts: 36 | Registered: 20 September 2006Reply With QuoteEdit or Delete MessageReport This Post
<ladytech>
Posted
WinkI have posted this before, Place the mach in rinse allow the mach to start counting down and then raise the shunt door look carefully for a leak at the Hansen Coonecters, have found the o-ring or the whole connecter to fix the problem. This WORKS!!! for checking the Hansen Connecters for leaks. Razzer
 
Reply With QuoteEdit or Delete MessageReport This Post