Go
New
Find
Notify
Tools
Reply
  
Low flow error on T Machine
 Login/Join
 
<Guest>
posted
Need help with a 2008 T machine.
In rinse mode gives me a inlet flow error.
Machines does go into Dialysis mode and passes all test
But as soon as I go into rinse mode it's gives me this error.
I've replaced D&F motor brushes verified pressures.Swapped out V41/29-V30 drain
Also verified incoming water pressure.
Any help will be greatly appreciated
 
Reply With QuoteReport This Post
<Cubby>
posted
Check v39. It opens during disinfect cycles, but stays closed in dialysis mode.
 
Reply With QuoteReport This Post
<Guest>
posted
Thanks Cubby
 
Reply With QuoteReport This Post
posted Hide Post
Does it happen at the 14:51 mark. If so it could be board related or cable related or some kind of connection problem.
 
Posts: 5 | Registered: 19 July 2017Reply With QuoteReport This Post
<rebel ron>
posted
had one before do the same thing it ended up a bad reed switch for me.
 
Reply With QuoteReport This Post
<Guest>
posted
Make sure someone didn't turn on dialysis te sample in hardwire options. It will guess this exact scenerio.
 
Reply With QuoteReport This Post
posted Hide Post
Change CFS # 10.


These opinions are my own and not of my employer.
 
Posts: 708 | Location: Buffalo, NY | Registered: 14 April 2012Reply With QuoteReport This Post
<Guest>
posted
V39 was the culprit.
Thanks for the input.
 
Reply With QuoteReport This Post
  Powered by Social Strata  
 