<Robb>
Posted
I have a FMC 2008K thats been problematic for a month now. It all started with wacky conductivity and an intermittent low flow error. I found this to be a bad balancing chamber membrain. I did not have any new ones but found some in an older balancing chamber that I had lying around. After putting the membrains in, the problem went away for a couple of days. Now the problem is worse I have low flow errors cycling with high flow errors and the conductivity is either 17 or at 10. But these problems only happen when a pt. is on. I can run this thing in the shop all day long and no alarms at all. Any help out there?
 
Reply With QuoteEdit or Delete MessageReport This Post
<bio>
Posted
Need to but membrains in check flow and degras pressures depending how many hr on mach my need new brushes in both pumps.
 
Reply With QuoteEdit or Delete MessageReport This Post
<bio>
Posted
quote:
Originally posted by bio:
Need to put new membrains in check flow and degras pressures depending how many hr on mach my need new brushes in both pumps.
Worded somethings wrong
 
Reply With QuoteEdit or Delete MessageReport This Post
<Robb>
Posted
I've already replaced the flow motor and pump head. I've calibrated the deair acid bicarb and uf pumps. I tried to calibrate the balancing chamber and the fluid discharge was 70mil. the machine wouldn't let me enter the number because it was too high. the highest setting is 63mil.
 
Reply With QuoteEdit or Delete MessageReport This Post
<1234>
Posted
you need new membranes
 
Reply With QuoteEdit or Delete MessageReport This Post
<Robb>
Posted
Thanks man, I'll try that tomorrow But I still don't understand why it runs fine in the shop but the minute I put a pt. on the thing goes haywire. I even ran a mock treatment for 3hrs and It didn't alarm conductivity 1 time...
 
Reply With QuoteEdit or Delete MessageReport This Post
<bio>
Posted
Sometimes that just the way it is happens like that everwhere. Could b something getting trash in in hydro block. How many hr on mach?
 
Reply With QuoteEdit or Delete MessageReport This Post
<smoothy>
Posted
I ran into a similar problem. I found valves 36 and 38 to be bad. My machine had about 16,000 hour on it.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Robb>
Posted
This machine has roughly 18000 hrs. on it. I replaced the whole balancing chamber with one from another machine and the low/high flow errors went away. Now the damn conductivity wont go above 13.3. I re-calibrated De-air flow acid bicarb and uf pumps, checked loading pressure as well. any other suggestions?
 
Reply With QuoteEdit or Delete MessageReport This Post
<John>
Posted
Robb,

This sounds more like a valve on the BC intermittently failing. Try and monitor your valve output(s).

-John
 
Reply With QuoteEdit or Delete MessageReport This Post
<John>
Posted
Didn't really read your last post very well -

Re-check your membranes to make sure they didn't slide a little. A torn or uneven membrane will cause this problem. Also could be a leak? Do a valve leak test. Do a BC volume test with your new BC assy.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Robb>
Posted
I already measured the fluid volume of the b.c. it was almost right on 59 mil. will try a valve leak test thanks.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Peter>
Posted
Hello all, I am asking this question in this area because we have never changed a membrane in the balance chamber. In fact I have checked with Fresenius Australia and no one has ever ordered one. Our machines are the 4008 and 5008 series units, fully made in Germany and I have had machine with 32000 hrs with no problem, Fresenius say the avg life is 35000 hrs. Our machines are heat disinfected 3 times a day and bleached once a week with 5% bleach and never seen a problem with a membrane. Can you explain why you change membranes, what causes them to fail in your 2008 series and not in the 4008 series, the BC is the same part. Many Thanks
 
Reply With QuoteEdit or Delete MessageReport This Post
<Robb>
Posted
I think that the difference is that we bleach our machines every other day. We alternate between heat and bleach. I don't know if the membrain part number is the same but I do know that the bleach is what ruins the integrety of the rubber membrains.
 
Reply With QuoteEdit or Delete MessageReport This Post
<old F Guy>
Posted
quote:
Originally posted by Peter:
Hello all, I am asking this question in this area because we have never changed a membrane in the balance chamber. In fact I have checked with Fresenius Australia and no one has ever ordered one. Our machines are the 4008 and 5008 series units, fully made in Germany and I have had machine with 32000 hrs with no problem, Fresenius say the avg life is 35000 hrs. Our machines are heat disinfected 3 times a day and bleached once a week with 5% bleach and never seen a problem with a membrane. Can you explain why you change membranes, what causes them to fail in your 2008 series and not in the 4008 series, the BC is the same part. Many Thanks


Been working on H and K machines for 15 +yrs & have never had to change a bc membrane either. Have machines now in service with 30,000hrs + with the original membranes. Heat disinfect daily and bleach once a week.
 
Reply With QuoteEdit or Delete MessageReport This Post