jdbiomed - Thanks for the plug.
The plumber is correct; your system was installed per code. Unfortunately, code does not require it to be installed per AAMI. One way I have found to combat this is to put the AAMI instructions on the blue prints...probably too late for you. I have scared a few plumbers away by saying they are not FDA approved. Some union plumbers can be very difficult. They think that as long as they have a license, then they can do the job. Many of the dialysis water companies don't use licensed plumbers but they have been trained on AAMI. There are probably a few hundred clinics (if not thousands) in the same situation as you.
AAMI RD52 section 7.1 gives the best description of how a loop should be constructed to mimize bacteria proliferation.
90 degree elbow are not a big problem. They will cause a larger pressure drop across your loop, slowing down the water velocity. This can impact the AAMI requriement for a minimum 3 ft/sec flow velocity (indirect feed) or 1.5 ft/sec (direct feed) (AAMI RD52).
RD52 also states "Dead-end pipes and unused branches and taps that can trap fluid should be eliminated because they act as reservoirs of bacteria and are capable of continuously inoculating the entire volume of the system." I have defined a dead-leg as anything longer than 1.5 times the pipe ID. Every station will probably have a dead-leg. The hose to the dialysis machine is a dead-leg when the machine is not in use. To combat this, you may need to run a brush through the piping at each station. I have gotten long brushes from a pet store in the aquarium section. I would clean the inside of this piping with the brush every month during your monthly disinfection.
Also, "Joints between sections of piping and between piping and fittings should be formed in a manner that minimizes the formation of crevices and other voids that may serve as sites for bacterial colonization." "Pipes should not be cut with a hacksaw." "Any burrs should be removed before the joint is formed." And, "water distribution systems should be configured as a continuous loop and designed to minimize bacterial proliferation and biofilm formation" All of these quotes are from AAMI RD52.
The purpose of these statements in AAMI is to help minimize bacteria and biofilm formation. Biofilm forms in the dead-legs, crevices left inside the joints, collect on the burrs left on the piping, etc. The other way to control bacteria and biofilm is with a good disinfection routine (at least monthly).
The other option is to install endotoxin filters at each station to better protect your patients. If you are using the Fres K machine, you can use their Diasafe filter.
Call me if you need further information.
The Water Guy - Florian Services