I have a machine that is giving restriction from the acid pump into the Df sub rack causing my line from the pump to blow. I have checked all the valves in tms mode and all are opening and closing I checked RVFPE to see if it was stuck and it was ok..any suggestions would be greatly appreciated.
 
Report This Post
<Bunker 22>
posted
Check the line that goes to the DF block from the acid pump. Also, check RVK. On the machine at the top of the block is a grey cap. take it off and pull out the spring and disc. With the machine off, hook up a syringe to the input of the block from the pump and flush that cavity out.
 
Report This Post
<KND>
posted
Thanks Bunker that's what it was RVK
 
Report This Post
<Bunker 22>
posted
KND - Great job! Glad you got it fixed! Cool
 