Can someone please tell me how to manually put the machine through the online PHT test. I have a 2008k machine. Thanks in advance.
<Guest>
Posted
Ok, I got the machine to manually go into online PHT test. It passed the first couple times, as well as both alarm and pressure test and diasafe test. I have already replaced the diasafe because it was due 10/12. Dear pressure is -24, loading is +24. Now I have an intermintent high flow error, and the third time it did fail online PHT. On the valve test, all valves passes. Any suggestions????
<Guest>
Posted
It is a fresenius 2208k machine with 11259 hours
<Guest>
Posted
oops 2008k
<Guest>
Posted
If anybody has any suggestions, I would be very grateful. I plan on working on the machine tomorrow. Also temp and coonductivity have been calibrated with no problems
<sdct>
Posted
Online PHT failures are associated with valves 24, 25, and 26
It is one of those three valves that is failing.
<adam>
Posted
I just fixed my pht problem and could visibly see it was valve 25 leaking. easy to notice for me but i would check those 3 valves first.. good luck
<Northwest>
Posted
Go to advanced search and type in OLPHT (weblink below)