<dialysisgirl>
Posted
Need help. I have a phoenix machine that I can't get the ph meter to cal. I have swapped out ph board, ph probe and protective board with a known working one. I have used stablent 22 on the J18 and checked all voltages on the mother board.

Thanks in advance
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
My first suspect would be a bad probe. If you pulled a know good one out of another machine and it doesn't work, I would try taking the probe from the bad one and try it in the good machine. This is a good one, I'm not sure what my next step would be except to contact Gambro at 800-525-2623, they are usually very helpful. Please let me know what you find.
 
Posts: 91 | Registered: 16 September 2009Reply With QuoteEdit or Delete MessageReport This Post