On March 7th I had a post describing an air leak problem I was having between the top of the hydroblock and the inlet of the deaeration pump. The final solution to this problem was to replace the hydroblock assembly. I had tried two used hydroblocks we had at our facility. Unfortunately they had the same problem. Once the new hydroblock was installed, there was no more air leak.
<ladytech>
Posted
Possibly could have replaced the seal on those hydroblocks. Would have been a lot cheaper!!
<Lumpy>
Posted
I did try replacing the hydroblock seal first. It made no difference. Fresenius suggested we replace the entire hydroblock assembly.