One of my 2008h unit with a conductivity stays at 13.3, check acid & bi-carb pumps, ballancing chamber all ok. help is very much needed. thanks
 
Posts: 8 | Location: as.96799 | Registered: 14 June 2002Reply With QuoteEdit or Delete MessageReport This Post
<renalman>
Posted
I have experienced this problem a couple of times in the past. Even though I had a good BC volume of 60ml, the diaphrams would be the the cause of my problem. So, speaking from past experiences, I recommend replacing them if you haven't done so already. They may have a small pin hole or be streched.

Post again if this doesn't help you.

Renalman
 
Reply With QuoteEdit or Delete MessageReport This Post
<gizzman>
Posted
Don't be to hasty in concluding that its the chamber membranes. Did you calibrate the conductivity cell or cells if you have online clearence upgrade. If the membrane had a pin hole you should see somem pressure failures during test.

Renalman[/B][/QUOTE]
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Yes, I have tried cond. cell calibration and it stables at 13.3. and we dont have any units with OLC. Maybe Ill try the chamber. Thanks
 
Posts: 8 | Location: as.96799 | Registered: 14 June 2002Reply With QuoteEdit or Delete MessageReport This Post