Moderators: Dennis Todaro
Go
New
Find
Notify
Tools
Reply
  
-star Rating Rate It!  Login/Join 
<ghaa>
Posted
Yes we still have some 2008-D machines. I have one of these D's that whe the dialysate flow is turned up to 800 it will sometimes drop the conductivity to below 13.0. Then I turn the flow to 500 and the conductivity comes back up. It only seems to happen after the machine has been running for several hours. I checked the conductivity with an external meter and it really is going low. Please help.
 
Reply With QuoteEdit or Delete MessageReport This Post
<alex 2>
Posted
try replacing the balancing chamber membranes.
 
Reply With QuoteEdit or Delete MessageReport This Post
<ghaa>
Posted
Thanks for the help. It was the acid pump after it warmed up it would pump erratic.
 
Reply With QuoteEdit or Delete MessageReport This Post
 Previous Topic | Next Topic powered by eve community  
 


Copyright RenalWEB 2008