<Lumpy>
Posted
Hello Everyone :

I have a 2008K machine that peaks the conductivity to 16.0 during OLC test and results in a high conductivity alarm. I replaced the actuator board two weeks ago. It ran fine for two weeks and now the problem has returned. Any suggestions would be appreciated.

Thanks
Lumpy
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Try changing the whole acid pump. Putting new springs and poppets does not work.
 
Posts: 12 | Location: DOWN SOUTH | Registered: 05 May 2005Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I assume you calibrated the conductivity cells?
 
Posts: 205 | Location: Mountain View, CA | Registered: 19 October 2004Reply With QuoteEdit or Delete MessageReport This Post
<Tech>
Posted
calibrate A and B pumps and condo cells.
 
Reply With QuoteEdit or Delete MessageReport This Post