At our unit many patients have plummeting CO2 levels. Its been suggested that changing to Naturlyte 9000 bathe could result in a mix of 40 for BiCarb bath. If done, what adjustments would have to be made to the machine and acid mix, other?
 
Posts: 130 | Registered: 19 April 2000Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
That is what we were using. We are now using Rockwell solutions that yeld a 39 co2. The only thing that we had to change was the formula in the service mode. The add/enter concentrates screen. Switched from fresenius to drake and recailbrated the conductivity cells. To my knowledge this is all that was done to the machines. I was on vacation at the time as was not here when the switch was made. I will do some checking and get back with you.

Matt
 
Posts: 124 | Location: ardmore,ok | Registered: 08 December 2002Reply With QuoteEdit or Delete MessageReport This Post