Please help. I've got several machines that are continually being pulled for TMP problems. The techs say that during recirc. they alarm constantly & cannot be reset OR once the patient gets hooked up they start having TMP alarms. Naturally, I can hook the machines up in my work area & run them for hours without one single alarm.
TMP the scurge of fresenius dialysis machines! There are so many things that can cause a TMP error I am not really sure where to begin. It can be simple like a hansen not being pushed on hard enough to a problem with a stick. I guess I would suggest checking your Venous Arterial and Dialysate proceedures in the preventive maintenance manual. If any of these are off it will give you a starting point since you seem to be having multiple problems.
Posts: 123 | Location: Central Texas | Registered: 10 July 2007
Just test and disinfect machine and return to service. The problem is the access.
<rob>
Posted
Do you have diasafe filters on these machines?
<BayAreaBio>
Posted
It can be several things. It can be wet transdusers, clogged dialyzer or like s.aug said there access. Or they might be running a puff and the are plugged in to central acid or bicarb.
<Chem. Memo>
Posted
sorry did that problem occur with many dialyzers or with one dialyzer? i meet this problem before and by inspection we found that Ultrafiltration rate of Dialyzer is extremely Low and lose it`s ability to remove desired water. please check this point.