I have gotten an A,24 error on the blood pump on a K with 5k hours. Second time this came up in two months. Could not duplicate and dn't want to spend the money on a blood pump unless I have to.

Any suggestions to help verify?
 
Posts: 55 | Registered: 04 October 2007Reply With QuoteEdit or Delete MessageReport This Post
orc
Posted Hide Post
Try a blood pump speed calibration.
 
Posts: 360 | Registered: 20 January 2005Reply With QuoteEdit or Delete MessageReport This Post
<Atlanta Tech>
Posted
5K hours shouldn't have worn the brushes out on the motor, however, there is that confounded tach-o-meter.
If the calibration does not work and If you have an hour and an known working bloodpump module, swap the motors. Then run it in dialyze mode for a few minutes. Alarm code returns, then you have a board problem.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I checked the arterial pump cal and when I set the 100 RPMs, it was originally run 5 RPMs faster. The cal took and I am letting it run. I never could get it to mess up for me before, so it may be a long time before it fails again.
 
Posts: 55 | Registered: 04 October 2007Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Failed again, bought new pump motor and installed. Tested fine. Put it back into service.
 
Posts: 55 | Registered: 04 October 2007Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Send the bad motor to H&S Technical for rebuild. And did you remember to remove the O ring and bushings from the old motor and put them on the new motor? I have seen several of those bad motors around that still had the O ring on the nose and the bushings for the mounting screws still in them. Fresenius does not provide those when they send out a new motor, and it's really easy to overlook them.
 
Posts: 445 | Location: Cleveland.OH | Registered: 03 February 2006Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I changed it out and compared the two pumps to change over any parts. I think I got it all unless something was not visually obvious. It ran fine and passed my tests. Anyway, I held on to the old one just in case that was nto the solution.
 
Posts: 55 | Registered: 04 October 2007Reply With QuoteEdit or Delete MessageReport This Post
<brt>
Posted
Check the connection from the pump to the board, I've found it to be a big issue with A 24 alarm.
 
Reply With QuoteEdit or Delete MessageReport This Post
<Dr. Machine>
Posted
Have you had an issue with this pump chettrick since you replaced the motor? I have one doing the exact same thing, and after checking connections, recalibrating the speed, its still doing giving the alarm.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Dr Machine
IT has been awhile but I have not problems since, that I remember.
 
Posts: 55 | Registered: 04 October 2007Reply With QuoteEdit or Delete MessageReport This Post
<Dr. Machine>
Posted
Oh jeez, sorry man, I didn't realize this was an OLD thread, haha...someone must have bumped it. Yea, I just ordered the new motor for it, since the Tachometer is attached to the frickin' thing, and can't be removed. Thanks for the reply
 
Reply With QuoteEdit or Delete MessageReport This Post