I have noticied that Fresenius changed the bicard/Acid probe filters in the last 6 months or so. The problem is that I am all of a sudden getting black / greyish material in the filters and it's plugging them up in a matter of two months. I have never had this problem before. I am having this happen in all 5 of my units. Any ideas???
 
Reply With QuoteReport This Post
Posted Hide Post
I the past I've seen something like this happen. It turned out to be the type of vinegar I was using for the acid rinses. Some of the cheaper vinegar has sediment in the bottom and this was plugging up the filter on the bicarb line.
 
Posts: 67 | Location: Ohio | Registered: 15 February 2000Reply With QuoteReport This Post
<ghaa>
Posted
Thanks for the help, the vinegar was the problem.
 