Have a machine with high temp issues. It pegs at 44 then goes back down to around 36.5 then just keeps cycling like that and yes it is matched on my meter, while cond stays stable at 13.8 even though thats probably not to accurate cause of the high temp. I have replaced senor board w/ cable, NTC2, Function board, heater rod, and the triac. Any help would be greatly appreciated.
<Guest>
Posted
Could it be your flow? I would check dearation and flow pressures.
<Techmex>
Posted
If no flow errors, PWR Control Bd could be ramping the voltage to the HTR element.
<brreed>
Posted
It ended up being the power supply, however now that I have temp at 37 and cond at 13.7 Im failing actuator temperature low. Have changed out actuator board/cable. sensor cable, both test boards, and motherboard with no change any ideas?
<sdct>
Posted
Calibrate the temp sensors and heat temp control
<sdct>
Posted
one more thing. The problem was most likely the power control board. The rest of your power supply is most likely good
<bio>
Posted
Had the same problems last week, problem was thr Triac, you didnt have to buy the whole power supply
<brreed>
Posted
Tried cal the temp sonsors but get operator error when I do post temp.
I'm having the same problem except I changed out the Power Supply, all the cards, the heater rod, ntc-2. and the cables to the distribution board and still no solution.