Having a problem with the Fresenius 2008 machine. The TMP is topping out and the acid and bicarb pumps are not running. I know there is a flow error but there is not alarm. The heater will not come on. I replaced the float switch. Checked the flow and deaeration pumps and they are good. Replaced valve 25 and 39. Please help.
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I would say with the multiple problems you are describing, I would look at the power supply or a circuit board, but here are some helpfull tips I would look at:

1: Flow error - insert the loading pressure gauge and see what pressure you have there - depending on the reading, depends on how to preceed. Call up debug screen #5 and check the water and flow error numbers - 0= no problem, 1= problem.

2: Temperture - easy to over look, but check to make sure the circuit breaker on the power supply is turned on - I have seen staff accidently flip it off. If it is on I have seen bad breakers or most of the time it is someting with the power supply causing the problem.

3: Aid and bicarb pumps not working - now I have not seen where both of them do not work at the same time. I would check to see if they might possibly be vapor locked. With the hydrolic section out you can feel if the pumps are working or not.

4: TMP - I have seen bad valves and the CFS switches cause this problem.

Hope this helps.


Working mans PH'D
 
Posts: 49 | Location: Midwest | Registered: 30 August 2007Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
INLet flow error is caused 90% of the time by the actuator board. I have seen the problem several times. I swapped mine out and it worked each time. H&S repaired my boards at a very reasonable price. I think it was like 125.
 
Posts: 7 | Location: I-R-AZ | Registered: 06 May 2008Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
The drivers for both the acid and bicarb pumps are also on the actuator board.
 