Moderators: Dennis Todaro
Go
New
Find
Notify
Tools
Reply
  
-star Rating Rate It!  Login/Join 
<nonameuno>
Posted
Does anyone know what would cause an A.24 blood pump code on a K machine? It happened at the end of my shift today. I could not stay waiting for Fresenius to call back.

Thanks.
 
Reply With QuoteEdit or Delete MessageReport This Post
<dibron>
Posted
Did you find out what was the problem? I got the same problem and the operators manual says
on that alarm "Optical Tachometer not in range".
(page128).
I ran machine with the lines being set up for 5 hrs- did not get it, but with the patient on it does after 2-3 hrs.Will call Fresenius, but what you found anyway?
Thank you
 
Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
I have had this problem before and the only thing that resolved it was to replace the motor tachometer assy.
 
Posts: 12 | Location: South Florida | Registered: 07 December 2005Reply With QuoteEdit or Delete MessageReport This Post
Posted Hide Post
Most likely motor assembly or 955
 
Posts: 6 | Registered: 29 December 2006Reply With QuoteEdit or Delete MessageReport This Post
<nonameuno>
Posted
It was the motor/tach assembly. The Fresenius tech said it was the 955 board or the motor. I had to run the machine for about four hours in the tech lab before the blood pump acted up again.

My employer has installed something called "Smart Filter" on all the networked PC's. It blocks me from viewing the Renal Web discussion boards at work.
 
Reply With QuoteEdit or Delete MessageReport This Post
<dibron>
Posted
Thank you guys! I am following the same steps as "nonameuno", but ordered 955 first as Fresenius advised due to the age of machine(1 yr) and 3000hrs only. They think it woud rather be a board...Will see.
 
Reply With QuoteEdit or Delete MessageReport This Post